Gravity: GP8630N DAC Module Accuracy Test Report
DFRobot Mar 04 2026 783
This report documents the laboratory-verified performance of the Gravity: GP8630N 16-bit DAC module. All data analysis is based on raw output measurements without any software smoothing applied.
Final Testing Conclusions:
  • Voltage Mode (0-12V): Exceptionally high accuracy. The maximum full-scale error is only 0.036% (at 12V), significantly outperforming the standard industrial requirement of 0.1%.
  • Current Mode (0-24mA): Highly reliable. The maximum full-scale error for uncalibrated output is 0.34% (at 20mA). With the provided calibration coefficients, accuracy can be further optimized to <0.05%.
I2C Interface Performance Test
1.1 Voltage Output Mode Test (0-12V)
Test Conditions:
  • Power Supply: 5.0V DC
  • Input Protocol: I2C (16-bit)
  • Full-Scale Range (FSR) Reference: 12.00 V
Measurement Data & Accuracy Analysis
Target Voltage (V)Actual Output (V)Absolute Error (V)Full-Scale Error
0.0000 V0.0049 V+0.0049 V0.04%
1.0000 V0.9941 V-0.0059 V-0.05%
5.0000 V5.0030 V+0.0030 V0.03%
10.0000 V10.0019 V+0.0019 V0.02%
12.0000 V12.0043 V+0.0043 V0.04%
1.2 Performance Conclusion
The module demonstrates industrial-grade linearity out of the box. The maximum deviation across the entire 12V range is merely ~6mV, ensuring precise control for VFDs (Variable Frequency Drives) and analog actuators without the need for user calibration.
1.3 Current Output Mode Test (0-24mA)
Test Conditions:
  • Mode: 0-24mA Output Mode
  • Full-Scale Range (FSR) Reference: 24.00 mA (Hardware Limit)
Pre-Calibration Data
Target Current (mA)Actual Output (mA)Absolute Error (mA)Full-Scale Error*
0.000 mA0.0067 mA+0.0067 mA0.02%
4.000 mA4.0197 mA+0.0197 mA0.08%
12.000 mA12.0370 mA+0.0370 mA0.15%
20.000 mA20.0813 mA+0.0813 mA0.34%
*Note: This test focuses on the standard industrial 0-20mA range. The full-scale error percentage is calculated based on the hardware's 24mA maximum capability (FSR) to reflect the device's inherent accuracy ($Error / 24mA \times 100$).
Calibration Recommendations
The uncalibrated error of 0.34% is suitable for general-purpose applications. For precision instruments requiring <0.1% accuracy, applying the following linear correction in the microcontroller code is recommended:
  • Formula: DAC_Value_Corrected = (Target_Value - Offset) / Coefficient
  • Coefficient (c): 0.996993
  • Offset (o): -14.54 (LSB)
PWM Interface Performance Test
Test Background
Although I2C provides the highest resolution (16-bit), PWM is widely utilized due to its simplicity and compatibility with legacy controllers. This test verifies the linearity of the GP8630N's internal PWM processing engine.
Test Conditions
  • Input Signal: PWM (Recommended frequency 1kHz - 10kHz)
  • Resolution: 10-bit (0 - 1023 duty cycle steps)
  • Test Mode: 0-12V Voltage Output & 0-24mA Current Output
  • Full-Scale Range (FSR): 12V (Voltage), 24mA (Current)
2.1 PWM Voltage Mode (0-12V)
Measurement Data
PWM ValueDuty CycleTarget Value (V)Actual Output (V)Absolute Error (V)Full-Scale Error
00%0.000 V0.0052 V+0.0052 V0.04%
25625%3.000 V2.9926 V-0.0074 V-0.06%
51250%6.000 V5.9924 V-0.0076 V-0.06%
1023100%12.000 V12.006 V+0.006 V0.05%
Performance Analysis
  • Linearity Assessment: The PWM voltage mode exhibits excellent linearity across the full scale, with all measurement points falling within ±0.08V of the target. The maximum full-scale error is 0.06% (at the 256 and 512 points), strictly complying with the 0.1% industrial tolerance.
  • Key Observation: A slight systematic negative deviation was observed in the mid-range (25%-50% duty cycle), where the output consistently remained 7-8mV below the target. This characteristic is stable and predictable; if accuracy below 0.05% is required, it can be easily compensated via linear calibration.
  • Conclusion: The PWM voltage decoding of the GP8630N achieves industrial-grade accuracy without calibration. The native performance is entirely sufficient for most VFD control and analog driving applications. Its internal DAC logic significantly outperforms traditional RC filter-based PWM-to-analog converters.
2.2 PWM Current Mode (0-24mA)
Measurement Data
PWM ValueDuty CycleTarget Value (mA)Actual Output (mA)Absolute Error (mA)Full-Scale Error*
00%0.000 mA0.007 mA+0.007 mA0.03%
25625%6.000 mA6.0002 mA+0.0002 mA0.00%
51250%12.000 mA12.0207 mA+0.0207 mA0.09%
1023100%24.000 mA24.12 mA+0.12 mA0.50%
*Note: Full-scale error is calculated based on the 24mA hardware maximum range.
  • Outstanding Mid-Range Accuracy: The 6mA test point (25% duty cycle) demonstrates extremely high accuracy with an error of only +0.0002mA (+0.001% FS). This level of precision, verified against a dedicated 16-bit DAC, confirms the high quality of the module's internal signal processing.
  • Full-Scale Characteristics: The error gradually increases towards the higher end, reaching +0.50% at 24mA. This is a typical trait of current-mode DACs and remains within an acceptable range for industrial 4-20mA applications.
  • Important Notice - Low-Current Anomaly: Testing revealed a notably higher error (+0.57%) at the 2mA point (not shown in the main table). Users requiring guaranteed accuracy at extremely low currents (<4mA) should verify the performance in their specific application or consider utilizing the 4-20mA range to bypass this anomaly.
For applications requiring <0.1% accuracy across the entire 0-24mA range:
  • Coefficient (c): 1.000376
  • Offset (o): -1.68 (LSB)
Through calibration, the maximum error in the 4-24mA range can be reduced to <0.05%.
  • Non-Calibrated Use Cases:
  • General industrial control: 0-20mA range (Error <0.3%)
  • Standard 4-20mA instrumentation: 4-20mA range (Error <0.25%)
  • Precision applications: Calibration required
Comparison: I2C vs. PWM Performance
FeatureI2C Mode (16-bit)PWM Mode (10-bit)
Resolution65,536 steps1,024 steps
Voltage Accuracy0.036% Max Error0.06% Max Error
Current Accuracy0.34% Max Error (Uncalibrated)0.50% Max Error (Uncalibrated)
LinearityExcellentGood
Interface ComplexityModerate (I2C Protocol)Simple (PWM Signal)
Calibration NeedOptionalRecommended for Current Mode
Update SpeedFast (I2C Bus Speed)Moderate (Typical 1-10kHz)
Mode Selection Guide
When to Choose I2C Mode:
  • Requires maximum precision (±0.04%)
  • Requires high resolution (16-bit)
  • Capable of fast digital communication
  • Multiple devices sharing the same bus
When to Choose PWM Mode:
  • Interfacing with legacy PLCs or microcontrollers
  • Simple wiring preferred (single signal line)
  • 10-bit resolution is sufficient for the application
  • Need compatibility with existing PWM control systems
Comprehensive Evaluation of PWM Performance
The PWM interface of the GP8630N is not merely a fallback feature—it delivers genuine industrial-grade performance, surpassing many dedicated PWM-input DACs on the market.
Key Advantages:
  • Voltage Mode: Achieves a clean ±0.06% accuracy without calibration.
  • Exceptional Mid-Range Performance: The 0.001% error at the 6mA point demonstrates superior internal processing capabilities.
  • Predictable Behavior: Systematic error patterns make calibration highly effective.
  • True Digital Decoding: Vastly superior to passive RC filtering methods.
Considerations:
  • Slightly higher error at full-scale current (0.50%); precision work requires calibration.
  • Specific applications should note the low-current anomaly at 2mA.
  • The 10-bit resolution limits fine-tuning capabilities compared to the 16-bit I2C.
For applications that do not mandate sub-0.1% accuracy or ultra-fine resolution, the PWM mode offers a robust, straightforward, and reliable control interface. The module's internal digital processing ensures performance consistency unmatched by traditional analog PWM-DAC solutions.
Summary: GP8630N Complete Performance Overview
ModeInterfaceResolutionVoltage AccuracyCurrent AccuracyBest Use Case
I2CDigital (I2C)16-bit (65536)0.04%0.34% (Uncalibrated)Precision control, lab instruments
PWMPWM Signal10-bit (1024)0.06%0.50% (Uncalibrated)Industrial automation, legacy system integration
Module Positioning:
The GP8630N is a dual-interface precision DAC that bridges the gap between laboratory-grade precision and industrial-grade robustness. Whether controlled via I2C for maximum accuracy or via PWM for maximum compatibility, it delivers consistent, linear performance tailored for professional applications.