Conductivity probe not entering calibration mode
Account cancelled 2019-09-18 19:05:25 9028 Views19 Replies I cannot get my (k=1) conductivity sensor to enter the calibration mode. Nothing happens when I put into "ENTER" or "enter" into the serial monitor (without the quotes" "). The monitor does not respond to any entered commands in the serial monitor. Any suggestions welcome.
2019-11-29 07:58:50 my probe is entering calibration, i can set it and save and exit. it will display + or- 3 ppm of what i set it to. but if i put it in another solution with a different ppm value i see no difference.
2019-11-29 07:58:50 my probe is entering calibration, i can set it and save and exit. it will display + or- 3 ppm of what i set it to. but if i put it in another solution with a different ppm value i see no difference.
2019-11-04 19:11:30 @iaglanda07
>>>Confirm Failed,Try Again<<< means that the calibration value is too different from the actual value, resulting in a failure to calibrate. Usually the value entered is not reasonable, or the probe is abnormal.
In addition, please note that in general, the calibration value is suitable between 1/2 and 2/3 of the measurement range.
This measurement range of this TDS sensor is 0~1000ppm, so the calibration value you used(that is, cal:5) is not suitable. It is too small.
You could use a large value to try again, like 707 ppm is better.
>>>Confirm Failed,Try Again<<< means that the calibration value is too different from the actual value, resulting in a failure to calibrate. Usually the value entered is not reasonable, or the probe is abnormal.
In addition, please note that in general, the calibration value is suitable between 1/2 and 2/3 of the measurement range.
This measurement range of this TDS sensor is 0~1000ppm, so the calibration value you used(that is, cal:5) is not suitable. It is too small.
You could use a large value to try again, like 707 ppm is better.
2019-11-04 19:11:30 @iaglanda07
>>>Confirm Failed,Try Again<<< means that the calibration value is too different from the actual value, resulting in a failure to calibrate. Usually the value entered is not reasonable, or the probe is abnormal.
In addition, please note that in general, the calibration value is suitable between 1/2 and 2/3 of the measurement range.
This measurement range of this TDS sensor is 0~1000ppm, so the calibration value you used(that is, cal:5) is not suitable. It is too small.
You could use a large value to try again, like 707 ppm is better.
>>>Confirm Failed,Try Again<<< means that the calibration value is too different from the actual value, resulting in a failure to calibrate. Usually the value entered is not reasonable, or the probe is abnormal.
In addition, please note that in general, the calibration value is suitable between 1/2 and 2/3 of the measurement range.
This measurement range of this TDS sensor is 0~1000ppm, so the calibration value you used(that is, cal:5) is not suitable. It is too small.
You could use a large value to try again, like 707 ppm is better.
2019-11-02 10:24:09 HI, i have got my SEN0244, and i´m trying to calibrate it using the advanced tutorial and the library on an Arduino Uno REV 3 with Arduinos IDE 1.8.8. I´m trying to calibrate but, when i put the command "cal: 5" only appears at the serial monitor the following message
">>>Confirm Failed,Try Again<<<"
I have cleared the EEPROM but it still doing the same
then, i have tried to put de command "calec:5 " but it still appearing the same answer, please help me. In addition. the "enter" command and the "exit" command seems to works well, cause the serial monitor response at them
">>>Confirm Failed,Try Again<<<"
I have cleared the EEPROM but it still doing the same
2019-11-02 10:24:09 HI, i have got my SEN0244, and i´m trying to calibrate it using the advanced tutorial and the library on an Arduino Uno REV 3 with Arduinos IDE 1.8.8. I´m trying to calibrate but, when i put the command "cal: 5" only appears at the serial monitor the following message
">>>Confirm Failed,Try Again<<<"
I have cleared the EEPROM but it still doing the same
then, i have tried to put de command "calec:5 " but it still appearing the same answer, please help me. In addition. the "enter" command and the "exit" command seems to works well, cause the serial monitor response at them
">>>Confirm Failed,Try Again<<<"
I have cleared the EEPROM but it still doing the same
2019-10-04 23:50:35 The correct commands has been shown as below:
ENTEREC -> enter into the calibration mode;
CALEC -> calibrate with one point when in the calibration mode;
EXITEC -> exit from the calibration mode;
ENTEREC -> enter into the calibration mode;
CALEC -> calibrate with one point when in the calibration mode;
EXITEC -> exit from the calibration mode;
2019-10-04 23:50:35 The correct commands has been shown as below:
ENTEREC -> enter into the calibration mode;
CALEC -> calibrate with one point when in the calibration mode;
EXITEC -> exit from the calibration mode;
ENTEREC -> enter into the calibration mode;
CALEC -> calibrate with one point when in the calibration mode;
EXITEC -> exit from the calibration mode;
2019-09-27 03:07:27 I have the same issue, I enter in "ENTER" in the serial monitor and nothing happens.