How to fix the Gravity Analog pH Sensor?

userHead Rojen VincentMedino 2022-05-15 12:27:40 899 Views2 Replies

Good day everyone, How to fix the Gravity Analog pH Sensor  the output pH reading is always at 15.45 pH.

2022-05-16 00:00:13

Only when the transmitter board output 0V can it prints pH 15.45 on the serial monitor.

Please check whether the transmitter board gets the suitable supply voltage range, that is, 3.3~5V.

Then short circuit the BNC input pin, the pH value on the serial monitor should be about 7.0. If not, please contact [email protected] for further help.

userHeadPic Youyou
2022-05-15 12:33:21