Two-Point Calibration or Can I Use One-Point Calibration for the Electrical Conductivity Sensor

userHead Tonny12138 2024-12-04 17:24:21 252 Views1 Replies

During the calibration, is it necessary to complete two points calibration? Can I complete only one-point calibration for the Analog Electrical Conductivity Sensor K=10?

2024-12-31 17:26:32

The two points calibration must be completed because the two points respectively correspond to standard buffer solution 1413us/cm and 12.88ms/cm. The two points calibration ensure the accuracy of the entire measurement range. Therefore, two points calibration must be completed.  

userHeadPic Tonny12138