1. Short time (over night to a week ): dip the probe into purified water or deionized water to prevent the filling solution from evaporating. Disconnect the probe from the main instrument every time when it is not in use.
2. Long time: (over a week): unscrew the cap from the probe and wash the electrodes core (cathode: platinum, anode: lead) and wash the cap with purified water or deionized water. Dry all the components with tissue. Screw the cap back to the probe without adding any filling solution to prevent the anode from being consumed. Put all the components back into the package.