why the calibration is not started after the 100g object is placed

When calibrating a 100g object, why the calibration is not started after the 100g object is placed, and the calibration fails all the time.
2024-07-11 09:41:16 A "trigger weight" is required for calibration. Press the calibration button and the sensor will only start calibration when it detects that something is put on and exceeds the "trigger weight". When in the first use, because there is no calibration, the data read of the 100g object by sensor may only be 20g. At this time, you need to reduce the value of MyScale.setThreshold(30) (change 30 to 10 or lower), and then upload the code again.
Tonny12138